During calibration and ironconstantan thermocouple is


During calibration, and iron/constantan thermocouple is zeroed( set to emit a Zero Voltage) AT 0 DEGREES CELCIUS. At 750 degrees celcius, it emits a voltage of 38.8mV. Alinear Out/input relationships exists between 0 and 750 degrees celcius. Determine the tranfer function of the thermocouple and the temperature to a voltage output of 29.6 mV.